tommyboy14
Well-known member
Hi Team,
The PROSE trial, a randomised control trial between ON-X and St Jude valves, have now published their results.
From the abstract:
The total patient population (N = 855) included patients receiving an On-X valve (n = 462) and a St Jude Medical valve (n = 393). The overall freedom evaluation showed no differences at 5 years between the prostheses for thromboembolism or for valve thrombosis. There were also no differences in mortality.
From the text, they find lower rates of complications for both valves than previous studies:
The major late hemorrhagic rate for the On-X prosthesis was 1.0% per patient-year (n = 23) and for the St Jude Medical prosthesis was 1.2% per patient-year (n = 23). The major hemorrhagic rates were not differentiated by prostheses overall, by aortic and mitral valve positions, or by economic development. The TE event rates were undifferentiated for the On-X prosthesis at 0.5% per patient-year (n = 12) and for the St Jude Medical prosthesis at 0.5% per patient-year (n = 10)
The most prominent major complication was VT (10 events in 9 patients). Within the total population, the On-X prosthesis major complication rate was 0.2% per patient-year (n = 5) and St Jude Medical prosthesis major complication rate was 0.3% per patient-year (n = 5). The aortic valve position major complication rate was 0.1% per patient-year (n = 2) and the mitral valve position major complication rate was 0.5% per patient-year (n = 8) (P = .007). The thrombosis rate was differentiated by economic development: 0.04% per patient-year (n = 1) for the Western population versus 0.5% per patient-year (n = 9) in the Developing country population (P = .005). Review of anticoagulant therapy records in all VT patients showed that the INR status varied extensively or was not followed. The time postoperation from the original surgery was mostly relatively early (<1 year) but varied up to 4 years. One of the On-X prosthesis aortic position cases was not receiving anticoagulation therapy at all.
The PROSE trial, a randomised control trial between ON-X and St Jude valves, have now published their results.
From the abstract:
The total patient population (N = 855) included patients receiving an On-X valve (n = 462) and a St Jude Medical valve (n = 393). The overall freedom evaluation showed no differences at 5 years between the prostheses for thromboembolism or for valve thrombosis. There were also no differences in mortality.
From the text, they find lower rates of complications for both valves than previous studies:
The major late hemorrhagic rate for the On-X prosthesis was 1.0% per patient-year (n = 23) and for the St Jude Medical prosthesis was 1.2% per patient-year (n = 23). The major hemorrhagic rates were not differentiated by prostheses overall, by aortic and mitral valve positions, or by economic development. The TE event rates were undifferentiated for the On-X prosthesis at 0.5% per patient-year (n = 12) and for the St Jude Medical prosthesis at 0.5% per patient-year (n = 10)
The most prominent major complication was VT (10 events in 9 patients). Within the total population, the On-X prosthesis major complication rate was 0.2% per patient-year (n = 5) and St Jude Medical prosthesis major complication rate was 0.3% per patient-year (n = 5). The aortic valve position major complication rate was 0.1% per patient-year (n = 2) and the mitral valve position major complication rate was 0.5% per patient-year (n = 8) (P = .007). The thrombosis rate was differentiated by economic development: 0.04% per patient-year (n = 1) for the Western population versus 0.5% per patient-year (n = 9) in the Developing country population (P = .005). Review of anticoagulant therapy records in all VT patients showed that the INR status varied extensively or was not followed. The time postoperation from the original surgery was mostly relatively early (<1 year) but varied up to 4 years. One of the On-X prosthesis aortic position cases was not receiving anticoagulation therapy at all.