I've encountered what I'm considerng may be a crisis of confidence. Perhaps Scribe with a lancet can share his experiences here.
For more than a year, I've seen lab and hospital results more than a full point higher than those of my Coag-Sense PT1 and PT2 meters. I've used a few batches of Coag-Sense strips and trusted my meter over the lab results.
I just got some strips for the CoaguChek XS (two packs of 6, expiring 5/31/2020) and used them for comparison to the Coag-Sense results. I expected the results to be a near match. The XS strips are not part of the recall. They were in a sealed package and the boxes didn't appear to have been tampered with.
My Coag-Sense gave me a value of 3.1.
The XS gave a result of 3.9. I tested again, and got a 4.2. Although I realize that the XS can give a higher than accurate result, increasing in error as the INR increases, I'm concerned by the significant difference between the two meters. I'm considering that the labs (both from a large corporation and hospitals) may be closer to the CoaguChek XS, and their results may be within a 20% margin of error with the results of the XS.
I asked Coagusense about the difference between meter and hospital labs, and haven't heard from them for a week.
I just asked again. I'll see if they respond.
Your other posts today may have part of the answer hidden in them. About a year ago, I was told to take 8 hour Tylenol for pain as it was the only medicine that did not interact with Coumadin. I ended up in the Urgent Care facility of my health plan as my internist was 95% convinced I had a major bleed and a high INR but had seen Hematomas from blood clots and wished to make sure that it was a bleed so the corrective action would not be wrong. Right after seeing the Urgent Care doctor, getting an Ultrasound and a new INR, I was told to drop my Coumadin for a day and start a reduced dosage. Then, a different Coumadin Pharmacist called me and told me that the Coumadin Clinic had been wrong and that Tylenol can and will cause INR changes and I had to stay under 4 pills a day. I have subsequently found that my reaction to it varies and can sometimes be very strong and relatively short lived - 24 hours or less.
You mentioned that "yes, required dosing varies from individual to individual - it's often done by body weight - " We were told by a pharmacist that many medicines no longer carry corrections for dosages for age and body mass in their package inserts and so the doctors have stopped using those corrections. He also said that dose per ml of blood does vary by body mass and can and should have a real impact for most patients. 16 years ago, after my heart valve replacement, they used the then relatively new 8 hr tylenol to help wean me off of hydroxycodeine. The Tylenol did not affect my INR at all when I was younger. Nowadays, watch out. I am very glad that you and Pellicle recommended that I get my meter so that I can watch the reaction - and hopefully slip in an extra pill or two a day. Unfortunately, reactions to medicines do vary not only by individual but also by the age of that individual.
You mentioned in your other post that "... and, perhaps unknown to my cardiologist, another doctor prescribed an additional calcium channel blocker" This is why my health plan prefers that we get everything thru the same pharmacy chain and that the doctors make the order thru the same computer system. It forces BOTH the doctors and the pharmacists to look at all the medicines that are being prescribed and the computer system looks for conflicts.
Perhaps the glitches your doctors and their medical prescriptions are causing different kinds of glitches in the hospital lab results and the CoagUChek results then in the Coag-Sense results?
Since my Coag-Sense is giving me values consistently above the lab readings, I cannot comment directly on your readings. I found out that my health plans lab seems to have a 6 hour lag between when they draw the blood and when they result the INR. The ISI standard allows a drop of 10% in the INR for that time lag. If that is correct then my meter reading and the Lab reading would be always with 0.2 units and usually with 0.1 units. This differs from the regression analysis and correlation where only one reading was 0.2 units and the rest were within 0.1 unit.
This makes me confused about your high readings. Can you ask the hospital when was the last time they changed their reagent? Then ask them when was the last time they recalibrated their lab INR equipment and reagent against the ISI standard for INR measurement.
I have the advantage that I have only had the meter 9 months even though I have been on Coumadin 16 years. Thus, I have had the same lab, lab equipment, lab reagent, INR meter and meter reagent for that time. Even the same "lab" according to another poster, can send stuff out for processing elsewhere.
This whole business is trickier then I thought. I would like a way of better benchmarking the Coag-Sense but have found none better then your idea of having multiple meters and running them against each other. It now appears that to do that reliably, one must do it on a regular basis and I cannot currently afford that.
Walk in His Peace,