    Question Error Messages on INRatio

    The thing was working fine for a couple of months.
    today, I got three ERROR messages.
    QC1H, 411, and QC1 = 14.3 and QC2 =22.8

    The nice people at the support hotline told me that the Quality Controls were reading high, which anyone could figure out. Except, what does that mean, exactly?

    Does anyone know?
    They're sending more strips, but it concerns me.
    Anyone have similar experience?

    It's been hot here lately, but not over 90 degrees in my air-conditioned office, nor lower than 35 degrees.

    Any suggestions?

    [SIZE="5"][FONT="Comic Sans MS"]Temp69[/FONT][/SIZE]

    AVR replaced with On-X 6/27/06, bypass, Aortic Aneurysm repaired. UCLA, Dr. Laks

  2. #2
    Join Date
    Dec 2001
    Location
    On The Hot Seat
    Posts
    25,981

    Default

    Can you find somewhere cooler to store the strips and try again? I may be too hot or humid right were you are.

  3. #3
    geebee Guest

    Thumbs up

    Did all the error messages have the QC? My manual indicates a problem with the storage of the strips or the expiration date. Since I am sure you were using strips that have not expired, I agree with Ross - storage humidity or temp. Perhaps you should try refrigerating them (the butter section of the fridge is a good location since it is a little warmer that the rest of the fridge).

    If the 411 was alone under the error message, it indicates that not enough blood was used. I used to get a lot of 3 digit error codes until I got more experienced.

  4. #4
    Join Date
    Jun 2001
    Location
    McLean, VA
    Posts
    1,586

    Default

    I have not gotten any error messages since I made sure to have plenty of blood (>15 ml) with NO air bubbles in it. I use capillary tubes to get the blood off my finger and then squirt onto the strip.
    Marty
    Mitral valve replacement
    September 17, 1998
    Mechanical St. Jude
    Inova Fairfax, Falls Church VA
    Dr. Ed Lefrak

  5. #5
    Join Date
    May 2006
    Location
    calabasas, ca
    Posts
    419

    Default thanks for the advice. problems solved

    they shipped new strips, inr ok at 2.5
    was defective strips.

    will now store them in cooler place.
    but can't be below 32 degrees.